Concept:
A defibrillator is a life-saving medical device used to terminate lethal cardiac arrhythmias, such as ventricular fibrillation (VF), by delivering a controlled, high-energy electrical shock to the heart muscle. This massive electrical charge depolarizes a critical mass of the myocardium simultaneously, resetting the heart's electrical system and allowing the natural pacemaker (the Sinoatrial Node) to re-establish a normal sinus rhythm.
In medical electronics, the total electrical energy ($E$) discharged by a defibrillator capacitor is calculated in Joules ($\text{J}$), which is equivalent to Watt-seconds ($\text{W}\cdot\text{sec}$):
\[
E = \text{Power} \times \text{Time} = \text{Watts} \times \text{Seconds} = \text{W-sec}
\]
Step 1: Analyzing energy requirements for Implantable Cardioverter-Defibrillators (ICDs).
An ICD is surgically implanted inside the patient's body, with its lead wires placed in direct contact with the endocardium inside the heart chambers. Because the electrical shock is delivered directly to the heart tissue without passing through skin, fat, or the chest wall, there is no structural attenuation or current dispersion. Consequently, the energy required to fully depolarize the myocardium is low, typically falling within the range of $5 \text{ to } 30\text{ W-sec}$ (Joules).
Step 2: Analyzing energy requirements for External Transthoracic Defibrillators.
External transthoracic defibrillators deliver their shock non-invasively through large electrode paddles or adhesive patches placed on the patient's bare chest skin surface.
The electrical current must travel through the skin, subcutaneous fat layers, skeletal muscle, and ribs before reaching the heart. Because chest tissues present significant electrical impedance, a large portion of the energy is dissipated along the path.
Modern transthoracic units use optimized biphasic waveforms, which require substantially lower energy levels to achieve successful defibrillation compared to older monophasic designs. Standard clinical protocols for biphasic external defibrillation dictate an energy delivery range of $50 \text{ to } 100\text{ W-sec}$ for initial shocks (scaling up if necessary).
Step 3: Comparing units and ranges.
• Options (A) uses pure Watts ($\text{W}$), which is a unit of power rather than energy, making it scientifically incorrect.
• Options (C) and (D) present energy values ($500\text{-}1000\text{ J}$) that exceed standard biphasic clinical guidelines and could cause severe thermal myocardial tissue damage.
Thus, Option (B) correctly states both the proper units ($\text{W-sec}$) and the accurate energy delivery ranges.