An analog instrument has a specified accuracy of \(\pm 1%\) of full-scale reading. If its full-scale value is 300 V and it reads 120 V, the maximum percentage error with respect to the indicated reading is
Show Hint
The absolute error remains constant at 3 V across the entire scale. As the measured voltage reading decreases, this fixed 3 V uncertainty becomes a larger percentage of the reading: \(\frac{3}{120} \times 100 = 2.5%\).
Concept:
The static error limit of an instrument is often specified as a percentage of its Full-Scale Deflection (FSD) value. This fixed error value remains constant across the entire measurement range of the instrument scale:
\[
\text{Absolute Limiting Error } (\delta V) = \pm (\text{Accuracy Grade}) \times (\text{Full-Scale Value})
\]
When evaluating accuracy at an intermediate measurement point (the indicated reading), the relative limiting percentage error increases and is calculated as:
\[
% \text{ Relative Limiting Error} = \frac{\text{Absolute Limiting Error } (\delta V)}{\text{Actual Indicated Reading } (V_{\text{actual}})} \times 100%
\]
Step 1: Calculate the absolute limiting error value from the full-scale specifications.
Given parameters:
• Full-scale range value = 300 V
• Base full-scale accuracy limit error = \(\pm 1%\)
\[
\delta V = \frac{1}{100} \times 300\text{ V} = 3\text{ V}
\]
This means any reading taken on this scale has an inherent absolute uncertainty limit of \(\pm 3\text{ V}\).
Step 2: Calculate the relative percentage error at the indicated value.
The instrument displays an indicated reading value of 120 V. Using our relative limiting error formula:
\[
% \text{ Error at 120 V} = \frac{\delta V}{V_{\text{indicated}}} \times 100% = \frac{3}{120} \times 100%
\]
Simplifying the fraction:
\[
\frac{3}{120} = \frac{1}{40}
\]
Now substitute this value back into the percentage equation:
\[
% \text{ Error} = \frac{1}{40} \times 100% = \frac{10}{4}% = 2.5%
\]
Thus, the maximum relative percentage error at the indicated reading is \(2.5%\), matching option (C).